EKG Shield - signal path

Started by danvol23, February 24, 2020, 11:52:21 am

Previous topic - Next topic

danvol23

Hello,

I bought Olimex EKG Shield and I managed some measurement of the path which the signal goes through the board according to the schematic in the datasheet. It was done because it seems to me that the signal is still very noisy at the output. With the oscilloscope, I discovered that the signal firstly appears on the C9. How does the signal go for example through the instrumental amplifier or the gain after that when the R30 and R31 are missing. Through the V_REF? I do not really understand it. Should I add some components? The signal still has the baseline wander at the output as well. If you could describe to me this, I would be grateful.

Thank you,
Daniel

LubOlimex

February 24, 2020, 03:08:32 pm #1 Last Edit: February 24, 2020, 03:18:27 pm by LubOlimex
R30 and R31 are not missing, these are board mounted meaning there is connection (short-circuit between the pads).

This is done so that you can place additional low pass filter if you decide (using the pads for resistors R30 and R31 and C28 and C29 the capacitors - remember to cut the wire in R30 and R31 before solder placing anything over it).
Technical support and documentation manager at Olimex