Others => Biofeedback - EEG, ECG, EKG => Topic started by: Romain on November 12, 2013, 04:38:57 PM

Title: EEG-ANALOG-ASM active electrode compatibility
Post by: Romain on November 12, 2013, 04:38:57 PM
Are all channels of the EEG-ANALOG-ASM board compatible with both passive and active electrodes?

Romain
Title: Re: EEG-ANALOG-ASM active electrode compatibility
Post by: LubOlimex on November 12, 2013, 05:02:24 PM
Hello Romain,

DRL needs to be passive always.

IMPORTANT: The electrodes OLIMEX sells (both AE and PE) are not compatible with EEG-ANALOG-ASM!!! These electrodes are compatible only with EEG-SMT! You have to make own electrodes for the EEG-ANALOG-ASM board. Information how to do it might be found here: http://openeeg.sourceforge.net/doc/hw/electrodes/ (http://openeeg.sourceforge.net/doc/hw/electrodes/).

Best regards,
Lub/OLIMEX