DUINO => SHIELDS => Topic started by: nescivi on February 07, 2014, 07:07:44 PM

Title: Calibration of the EMG shield
Post by: nescivi on February 07, 2014, 07:07:44 PM
Hi,

I was wondering if there are some details on how to do the calibration of the EMG shield.
I assume the aim is to get a maximum response out of the amplification?

sincerely,
Marije Baalman
Title: Re: Calibration of the EMG shield
Post by: LubOlimex on February 10, 2014, 02:26:09 PM
Hello,

You can use the trimmer for calibration. Generally, the idea is to try to keep the signal curve as big as possible without any "cuts" of the curve occur. A "cut" is when the peak or the low point of a measured signal gets abruptly flat.

The fine calibration is performed if you feed the square calibration signal CAL (V_CAL; V_GND) to the phone jack. This signal was used while calibrating the device here in the factory before shipping.

Best regards,
Lub/OLIMEX